Lesson 2 of 3 · Interferometry, response and calibration
Resolve and calibrate thermal radiance
24 minute lesson · 3 checks along the way
Your goal for this lessonExplain separate limits from optical-path sampling, finite path length and imperfect calibration.
In this lesson 4 sections · 3 checks
Read & understand
Measure interference before reconstructing a spectrum
A Fourier transform spectrometer combines light that has traveled along different optical paths. As their difference changes, the detector records an interferogram. For one ideal monochromatic component with wavenumber σ, the varying signal contains cos(2πσx), where x is optical path difference in centimetres. Different wavenumbers produce different oscillation rates. The spectrum is inferred from those oscillations after accounting for the instrument and its calibration.
Optical path difference is not automatically the physical travel of a moving mirror; the relationship depends on the optical arrangement. Here x always means the path difference itself. The following simplified calculations describe a uniformly sampled, otherwise ideal interferogram. They are useful for distinguishing sampling and resolution, but they omit phase errors, detector effects and other features of a complete processor. FORUM’s end-to-end simulator explicitly distinguishes simulated observations, calibrated spectra and later geophysical retrievals.
Read & understand
Sampling and resolving power answer different questions
If interferogram samples are spaced by Δx, an unaliased baseband requires wavenumbers below 1/(2Δx). With Δx=0.0005 cm, that limit is 1,000 cm⁻¹. Components at 400 and 1,600 cm⁻¹ then give the same sampled cosine values: at every integer sample the faster oscillation differs by whole cycles and a sign reversal inside the cosine. The samples alone cannot distinguish them. Choosing Δx=0.00025 cm raises the limit to 2,000 cm⁻¹ in this ideal example.
Maximum path difference sets another limit. Truncating the ideal interferogram at ±M produces a normalized line response sinc(2MΔσ), where sinc(z)=sin(πz)/(πz). For M=1.25 cm, the first zero lies 0.4 cm⁻¹ from the line center and the full width at half maximum is about 0.48268 cm⁻¹. These are different width definitions. Halving M doubles the response width. Merely printing a denser set of spectral points cannot undo a short measured path.
Check 1 of 3
10 pointsUse the reading above. You can try again without losing points.
Read & understand
Control the response and tie counts to radiance
A rectangular truncation produces sidelobes around a narrow feature. Apodization applies a smoother weighting to the measured interferogram, changing the line shape and usually trading lower sidelobes for a broader central response. It also affects how noise in neighboring spectral samples is correlated. A finer output grid is therefore neither proof of finer resolution nor proof of more independent measurements. Comparisons must use compatible response functions and uncertainty descriptions.
For a separate calibration exercise, assume detector counts C=gL+b with constant gain g and offset b. Two known reference radiances, L=2 and L=8 in a declared common unit, yield C=110 and C=410. The difference gives g=300/6=50; substitution gives b=10. A scene count of 260 therefore corresponds to L=(260−10)/50=5. Two references constrain both parameters in this ideal linear model. One reference alone cannot determine two unknown parameters without another supported constraint.
Finite path, finite resolution
Original ideal rectangular-window responses and a separate linear two-reference calibration exercise. Output sample density does not restore unmeasured path information. Values do not specify FORUM flight performance.
Open full-size diagram in a new tab: Finite path, finite resolutionCheck 2 of 3
10 pointsUse the reading above. You can try again without losing points.
Read & understand
A repeatable result can share a systematic error
Suppose both reference radiances are assigned values 0.2 too high while the same counts are observed. Their separation stays six, so the fitted gain remains 50. The fitted offset changes, and the scene is reported as 5.2 instead of 5. Repeating or averaging that biased calibration does not remove the common 0.2 error. Calibration requires an uncertainty chain for the reference itself, as well as a model of detector behavior over the relevant signal range.
ESA’s linked PTB photograph documents ground calibration hardware for an onboard blackbody. It is useful evidence of a calibration activity, not a measurement of achieved orbital retrieval accuracy. Real processing must also address spectral calibration, nonlinearity and response changes. A calibrated radiance product is an essential intermediate result; it is not yet a unique temperature, emissivity or cloud property. That distinction leads to the final lesson.
Check 3 of 3
10 pointsUse the reading above. You can try again without losing points.
Put it together
Your lesson progress
Each check connects to a part of the story. Complete all 3 to earn your 25-point lesson bonus.
Explore further · Optional
Go to the source
Explore the evidence behind this lesson. These optional readings and videos open in a new tab.
- Read: Ridolfi and colleagues: FORUM and IASI-NG synergyReviewed pages 1–3 and 5–6 discuss instrumental response and simulated retrieval. The lesson’s ideal OPD numbers are original exercises.
- Read: Sgheri and colleagues: processing through the simulatorReviewed pages 1–3 distinguish observation simulation, calibration and retrieval.
- Read: ESA: FORUM blackbody calibration at PTBGround hardware photograph credited to PTB; linked externally, without copying the third-party image into the site.